August 25, 2023 at 11:01 pm
Q:
System suitability in UV Methods
What criteria can be defined for system suitability in UV Methods for Assay, Impurities (Limit test)
Tagged: fghfg
All Replies
All Replies
Viewing 4 reply threads
-
Pharmacopeia do not define specific system suitability tests for UV methods. However the following may be considered.
- A test of repeatability by measurement of response of a standard solution, for example response of standard in five replicates (with %RSD to be less than 2.0%)
- Resolution with a solution of toluene in hexane, criteria – ratio between the minimum and maximum absorbance at respective wavelengths of 267 and 269 nm must exceed 1.3 (USP).
Useful references: