Q:

System suitability in UV Methods

What criteria can be defined for system suitability in UV Methods for Assay, Impurities (Limit test)

Methods, Validations

All Replies

Up
0
Down

Pharmacopeia do not define specific system suitability tests for UV methods. However the following may be considered.

  1. A test of repeatability by measurement of response of a standard solution, for example response of standard in five replicates (with %RSD to be less than 2.0%)
  2. Resolution with a solution of toluene in hexane,  criteria – ratio between the minimum and maximum absorbance at respective wavelengths of 267 and 269 nm must exceed 1.3 (USP).